Hello,
i'm wanted to test if the BCI2000 System can read our BrainVision RDA packets. Is there a way to only start the the source module and run the channel visualizer without the need of a application or signalprocessing module / configuration?
I managed to get a configuration file for the RDACLIENT using this tool. It seems correct. But i have no idea how to construct fitting signalprocessing / application modules. And i want to test the RDAClient before i get into this stuff.
i'm wanted to test if the BCI2000 System can read our BrainVision RDA packets. Is there a way to only start the the source module and run the channel visualizer without the need of a application or signalprocessing module / configuration?
I managed to get a configuration file for the RDACLIENT using this tool. It seems correct. But i have no idea how to construct fitting signalprocessing / application modules. And i want to test the RDAClient before i get into this stuff.
-
mellinger
- Posts: 1341
- Joined: 12 Feb 2003, 11:06
You may use any signal processing and application module if you just want to test whether signal acquisition works.
E.g., run the Stimulus Presentation configuration with the RDAClient source module, load the parms/examples/StimulusPresentation.prm parameter file, and click SetConfig. When the source module works, the operator module will display the source signal immediately after you press SetConfig, with the signal processing and application modules still inactive.
HTH,
Juergen
E.g., run the Stimulus Presentation configuration with the RDAClient source module, load the parms/examples/StimulusPresentation.prm parameter file, and click SetConfig. When the source module works, the operator module will display the source signal immediately after you press SetConfig, with the signal processing and application modules still inactive.
HTH,
Juergen
-
jkarch
- Posts: 4
- Joined: 29 Apr 2010, 11:12
I'm starting the following bat file:
Than i load
\parms\examples\StimulusPresentation_SignalGenerator.prm
cause i can't find StimulusPresentation.prm in examples.
Than i load my Rda params file which look like this:
(i replaced the real recorder ip with my ip in this post)
If i press Set config than the follow happens:
System Log:
The remaining behavior can be seen on the attached screenshot.
http://userpage.fu-berlin.de/juka87/error_bci.jpg
On the right you see the RDA Client by Brainvision connected to the same server.
By the way if i quit operate.exe RDAClient.exe keeps running and i have to kill it by myself.
Thanks for the help
Code: Select all
cd ..\prog
start operat.exe
start RDAClient.exe 127.0.0.1
start P3SignalProcessing.exe 127.0.0.1
start StimulusPresentation.exe 127.0.0.1
cd ..
\parms\examples\StimulusPresentation_SignalGenerator.prm
cause i can't find StimulusPresentation.prm in examples.
Than i load my Rda params file which look like this:
(i replaced the real recorder ip with my ip in this post)
Code: Select all
RDA string HostName= myip getparams myip %
RDA int SourceCh= 65 getparams myip %
RDA floatlist SourceChOffset= 65 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 // getparams myip
RDA floatlist SourceChGain= 65 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 0.1 1 // getparams myip
RDA float SamplingRate= 1000 // getparams myip
RDA intlist TransmitChList= 64 1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32 33 34 35 36 37 38 39 40 41 42 43 44 45 46 47 48 49 50 51 52 53 54 55 56 57 58 59 60 61 62 63 64 // getparams myip
RDA matrix SpatialFilter= 64 64 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 0 1 // getparams myip
RDA int SourceMax= 300muV // getparams myip
RDA int SourceMin= -300muV // getparams myip
System Log:
Code: Select all
30.04.2010 10:59:00 - BCI2000 started
30.04.2010 10:59:00 - Waiting for configuration ...
30.04.2010 10:59:00 - Waiting for configuration ...
30.04.2010 10:59:00 - Waiting for configuration ...
30.04.2010 10:59:30 - Operator set configuration
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of StimulusDuration for stimulus 1 will be 368ms rather than 375ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMinDuration for stimulus 1 will be 368ms rather than 375ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMaxDuration for stimulus 1 will be 432ms rather than 437.5ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of StimulusDuration for stimulus 1 will be 88ms rather than 93.75ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMinDuration for stimulus 1 will be 368ms rather than 375ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMaxDuration for stimulus 1 will be 432ms rather than 437.5ms.
30.04.2010 10:59:30 - Signal Processing confirmed new parameters ...
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of StimulusDuration for stimulus 2 will be 184ms rather than 187.5ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMinDuration for stimulus 2 will be 368ms rather than 375ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMaxDuration for stimulus 2 will be 432ms rather than 437.5ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of StimulusDuration for stimulus 3 will be 184ms rather than 187.5ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMinDuration for stimulus 3 will be 368ms rather than 375ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMaxDuration for stimulus 3 will be 432ms rather than 437.5ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of StimulusDuration for stimulus 4 will be 56ms rather than 62.5ms.
30.04.2010 10:59:30 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMinDuration for stimulus 4 will be 368ms rather than 375ms.
30.04.2010 10:59:31 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMaxDuration for stimulus 4 will be 432ms rather than 437.5ms.
30.04.2010 10:59:31 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of StimulusDuration for stimulus 5 will be 120ms rather than 125ms.
30.04.2010 10:59:31 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMinDuration for stimulus 5 will be 368ms rather than 375ms.
30.04.2010 10:59:31 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMaxDuration for stimulus 5 will be 432ms rather than 437.5ms.
30.04.2010 10:59:31 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of StimulusDuration for stimulus 6 will be 248ms rather than 250ms.
30.04.2010 10:59:31 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMinDuration for stimulus 6 will be 368ms rather than 375ms.
30.04.2010 10:59:31 - Warning: StimulusPresentationTask::Preflight: Due to a sample block duration of 8ms, the actual value of ISIMaxDuration for stimulus 6 will be 432ms rather than 437.5ms.
30.04.2010 10:59:31 - User Application confirmed new parameters ...
http://userpage.fu-berlin.de/juka87/error_bci.jpg
On the right you see the RDA Client by Brainvision connected to the same server.
By the way if i quit operate.exe RDAClient.exe keeps running and i have to kill it by myself.
Thanks for the help
-
gschalk
- Posts: 615
- Joined: 28 Jan 2003, 12:37
Source display ...
Hi,
do you get the source display with the SignalGenerator source module?
Gerv
do you get the source display with the SignalGenerator source module?
Gerv
-
gschalk
- Posts: 615
- Joined: 28 Jan 2003, 12:37
Brainvision ...
Try setting Source->SampleBlockSize to 40 and Visualize->VisualizeSourceDecimation to 5.
Gerv
Gerv
-
nsal
- Posts: 1
- Joined: 08 Oct 2012, 08:06
Re: Configuration with an Brainvision System
Hi,
I am having some problems with the configuration of the BrainVision System and BCI2000.
So I am trying to use the RDAClient as Signal Source. With 8 or 32 channels I do not have any problems, if i press set config then I can see the raw signals and the selected Application of interest runs without any problem. However, with 64 channels the system seems to crash.
Based on the previous post I have also tried to set the >SampleBlockSize to 40 and >VisualizeSourceDecimation to 5. However this does not change anything.
Do you have any ideas on this problem?
Neda
I am having some problems with the configuration of the BrainVision System and BCI2000.
So I am trying to use the RDAClient as Signal Source. With 8 or 32 channels I do not have any problems, if i press set config then I can see the raw signals and the selected Application of interest runs without any problem. However, with 64 channels the system seems to crash.
Based on the previous post I have also tried to set the >SampleBlockSize to 40 and >VisualizeSourceDecimation to 5. However this does not change anything.
Do you have any ideas on this problem?
Neda
-
mellinger
- Posts: 1341
- Joined: 12 Feb 2003, 11:06
Re: Configuration with an Brainvision System
Hi,
the RDAgetparams tool automatically creates a unit spatial filter matrix for you. With 64 channels, this is quite a large matrix, and results in considerable CPU load.
Set the Filtering->SpatialFilterType parameter to "none" in order to disable spatial filtering, and make sure to use a release build of the signal processing module (e.g., from the latest binary BCI2000 distribution).
Regards,
Juergen
the RDAgetparams tool automatically creates a unit spatial filter matrix for you. With 64 channels, this is quite a large matrix, and results in considerable CPU load.
Set the Filtering->SpatialFilterType parameter to "none" in order to disable spatial filtering, and make sure to use a release build of the signal processing module (e.g., from the latest binary BCI2000 distribution).
Regards,
Juergen