mellinger wrote: 11 May 2024, 16:03
Are you supposed to connect the Reference (In -) of Group A to the Reference (In-) of the other 3 banks where I shorted all Inputs (V+), Reference (V-) and Ground together? Isnt the effect like shorting the V- and ground of channel 1 while using channel 1 in Group A since the ground of all 4 Groups are also shorted??
No. Instructions for shorting everything applied to the case when you wanted to check for noise.
Just connect every input (except channel 1) to its associated reference. Then, connect all grounds together (by software setting, or using leads). For all channel groups except the first, connect the reference to ground, and make sure there is no internal connection between references (software setting).
Finally, connect your signal generator's + output to channel 1's + input, and the signal generator's - output to channel group 1's reference. If your signal generator has a ground output, connect it to the amplifier grounds.
Hope you can include the warning above when setting the default to True in CommonReference in BCI2000 (which is Default to False or disconnected in the Common References at g.USBamp Demo software).
Connecting all References and Grounds and shorting all inputs have the effect of shorting reference and ground in channel 1. Instead of the differential say 9-6 = 3.. you get only 9 because the V- is grounded making it a single ended measurement (not differential anymore), that's why the amplitude got higher. If I didn't use a signal generator and know the output is 10uV instead of the 25uV output displayed (which is only measuring V+ to ground and not deducting V- which is grounded), I won't know the connection is wrong.
So by setting only CommonGround to True and CommonReference to False with all inputs shorted except channel 1. I got back my noisy waveform.
Now the BMA-200 beat it again. Last night I was kinda celebrating thinking I got something better than the BMA (which I was trying to improve for months), then found out today it was all just an illusion. Lol.
By setting the CommonReference to False using the g.USBamp. At least the FFT doesn't have the big 958Hz peak and noises.
In the following is when CommonGround and CommonReferences both set to True using g.USBamp but with no input shorted producing the 958Hz and noisier which seems to be related to connecting all references.
Right now if I set all CommonGround and CommonReferences to True and all inputs shorted, which practically shorts the V- and ground in channel 1. I can't get an FFT because the magnitude of the V+ only signal is so huge.
Btw.. the noises in the 1st FFT above with CommonGround true and CommonReference false and 15 channels input shorted is also present in the BMA-200 in the following BMA FFT. This means the noises are related to the Netech Signal Generator. This was why I wanted to be able to see the BMA in BCI2000 and thanks to the fixed SoundCardSource, It did that.
I have some questions.
1. To move a control in BCI2000. What is the minimum channel required, should it be 8 that's why True is default in CommonReferences? it lead me astray 2 times with the 958Hz peak before and the false V+ only amplitude now. So for those using only one channel. it should be set to False. But can one move a cursor with one channel only?
2. You said "You record from multiple channels, and then combine them linearly to obtain a single signal that in turn is used to control the cursor.". Do you have any figure or table, etc. what is the increase in SNR or amplitude (or other variables) for this combined 8 channel vs 1 channel. I want see comparisons between 8 noisy channel vs 1 clear channel whether it is possible a good amplifier with 1 good channel can equal the 8 noisy channel. Or whether my BMA-200 with clearer 1 channel can equal the 8-channel noisier g.USBamp.
Thank you.