Noisy Signal
Posted: 19 Mar 2010, 05:08
Yesterday I ran a demo of a BCI on an ICT-conference and I experienced a few problems there. The demonstration was given in a huge hall that was bustling with network activity, other computers, mobile phones and the like, which made our signal incredibly noisy.
Fortunately we were not the only one to experience this problem, a few of our direct neighbors also had problems with their EEG setups. We discussed the problem together and tried some different filters (low pass, high pass and notch filters) and found a setting that gave us better results.
However we were unable to set our programs to these frequencies (i.e. a low pass filter of 20 Hz.), which amounts to the following question. Why can't we set the values of signal source to all possible values?
We were using a BCI2000 setup with gTec's g.USBamp (v2.0) and the signal acquisition program supplied with BCI2000. Other groups were using Biosemi hardware and software.
Fortunately we were not the only one to experience this problem, a few of our direct neighbors also had problems with their EEG setups. We discussed the problem together and tried some different filters (low pass, high pass and notch filters) and found a setting that gave us better results.
However we were unable to set our programs to these frequencies (i.e. a low pass filter of 20 Hz.), which amounts to the following question. Why can't we set the values of signal source to all possible values?
We were using a BCI2000 setup with gTec's g.USBamp (v2.0) and the signal acquisition program supplied with BCI2000. Other groups were using Biosemi hardware and software.