Problems with SynAmp Neuroscan amplifier during calibration

Known Issues and Problems with BCI2000
Locked
paolobernardis
Posts: 40
Joined: 02 Jul 2010, 04:51

Problems with SynAmp Neuroscan amplifier during calibration

Post by paolobernardis » 19 Aug 2010, 06:56

After installing BCI2000 and configuring the StimulusPresentation task, we are facing a problem of calibration when trying to set up the Neuroscan Scan 4.3.1 software. Specifically, we launch the acquire tool and when we calibrate the amplifier we record a 5 microvolt signal instead of 1 microvolt.
Is it possibile to be a result of BCI2000 installation? Can this problem be fixed?

Thanks in advance,
Paolo :shock: :roll:

gschalk
Posts: 615
Joined: 28 Jan 2003, 12:37

Neuroscan

Post by gschalk » 19 Aug 2010, 10:31

Hi,

Are you using the neurogetparams tool to get calibration values from the amplifier, and then loading those into BCI2000?

How do you assess signal amplitudes? Online or offline with Matlab?

Gerv

paolobernardis
Posts: 40
Joined: 02 Jul 2010, 04:51

Re: Neuroscan

Post by paolobernardis » 19 Aug 2010, 10:58

gschalk wrote:Hi,

Are you using the neurogetparams tool to get calibration values from the amplifier, and then loading those into BCI2000?

How do you assess signal amplitudes? Online or offline with Matlab?

Gerv
Hi Gerv,

thank you for the help.
The problem happen when I try to calibrate the amplifier using the Scan 4.3.1 software, before to run BCI2000. The Acquire module shows a 5microV value instead of the correct 1 microV, which is the default value sent into the circuit by the system to calibrate the electrodes.

My idea was to start with the Scan software, calibrate, prepare a subject with the ElectroCap, connect the Cap to the amplifier, and then put the amplifier in server mode, and last run the BCI2000. Is that sequence right?

thanks,

p.

gschalk
Posts: 615
Joined: 28 Jan 2003, 12:37

Neuroscan

Post by gschalk » 20 Aug 2010, 07:14

Hi,

If you see incorrect values in Acquire, isn't this a Neuroscan problem? If there is a constant factor that can correct the amplitude values, you can simply manually change the SourceChGain parameter within BCI2000 (see the wiki on doc.bci2000.org for more information on that parameter) to correct it.

Gerv

paolobernardis
Posts: 40
Joined: 02 Jul 2010, 04:51

Re: Neuroscan

Post by paolobernardis » 20 Aug 2010, 07:32

Hi,
gschalk wrote:If you see incorrect values in Acquire, isn't this a Neuroscan problem?
well, that's correct. but since the problem with the calibration appeared only after the installation of BCI2000, I was trying to figure out whether one (improbable) cause could be the BCI2000. Probably not, because I removed BCI2000 and reinstalled Neuroscan software, but the calibration problem persist.
gschalk wrote: If there is a constant factor that can correct the amplitude values, you can simply manually change the SourceChGain parameter within BCI2000 (see the wiki on doc.bci2000.org for more information on that parameter) to correct it.
thanks, i'll try this solution.

p.

Locked

Who is online

Users browsing this forum: No registered users and 2 guests